NRF 5340 resets automatically in between device scanning

hi,

                  i am trying to connect  peripherials ( blood pressure, pulseox meter) with nrf5340 using uuid and mac, initially using uuid filter are used to scan the devices around and store the mac of the device. once the mac is stored we use address filter to  filter the device based on the mac address that was stored earlier.

once the mac address matches with the stored mac address  the connect with the device is established and gatt discovery is initialed and the handles are collected and data is received and subscribing to them. everything works as expected.  and we could get the vitals from the blood pressure and pulse oximeter.

but we face an unexpected controller reset all of sudden in between in a random pattern whenever the scan_start function is initialed. since we use more than one device we initial Scan_start to initiate the filers to scan for other devices this is done during the connected, disconnected  and once gatt_ discovery call back function is succeeded. the nrf controller gets reset automatically during the scan_start function at any on the given call back function. 

static void set_scan_filters(void)
{
	// scan filters are determined based on the current mode (pairing or connecting)
	// and based on which (if any) connections are currently active

	int err;

	// remove existing filters
	//printk("Resetting scan filters\n");
	bt_scan_filter_remove_all();


	if(hub_mode == HUB_MODE_PAIRING_MODE)
	{
		// in Pairing mode, we want to just filter by UUID
		printk("Setting new scan filters for pairing mode based on UUIDs and current active connections\n");

		if(p_blood_pres_conn == NULL)
		{
			// not currently connected to any BP sensor, so add that filter
			//printk("Setting scan filter for BP UUID\n");
			err = bt_scan_filter_add(BT_SCAN_FILTER_TYPE_UUID, BT_UUID_BPS);
		}
		if(p_pulse_ox_conn == NULL)
		{
			// not currently connected to any Pulse Ox sensor, so add that filter
			//printk("Setting scan filter for Pulse Ox UUID\n");
			err = bt_scan_filter_add(BT_SCAN_FILTER_TYPE_UUID, BT_UUID_PUL_SERVICE);
		}
		if(p_therm_conn == NULL)
		{
			// not currently connected to any Thermometer, so add that filter
			//printk("Setting scan filter for Thermometer UUID\n");
			err = bt_scan_filter_add(BT_SCAN_FILTER_TYPE_UUID, BT_UUID_TEMP_SERVICE);
		}

		err = bt_scan_filter_enable(BT_SCAN_UUID_FILTER, false);
		if (err)
		{
			printk("Filters cannot be turned on (err %d)\n", err);
		}

	}
	else if(hub_mode == HUB_MODE_CONNECT_MODE)
	{
		// in Connect mode, we want to connect only if we have been given a target address and are 
		// not currently connected to that type of device

		printk("Setting new scan filters for connecting mode based on target addresses and current active connections\n");

		if(p_blood_pres_conn == NULL)
		{
			// not currently connected to any BP sensor

			// check if we have a specific address to connect to
			if(have_target_blood_pres_address == true)
			{
				// Have a target address - set filter
				//printk("Setting scan filter for target BP address\n");
				bt_addr_le_t addr_le_bps;
				memset(&addr_le_bps, 0x00, sizeof(addr_le_bps));
				memcpy(addr_le_bps.a.val, target_blood_pres_address, sizeof(target_blood_pres_address));
				addr_le_bps.type = BT_ADDR_LE_PUBLIC;
				err = bt_scan_filter_add(BT_SCAN_FILTER_TYPE_ADDR, &addr_le_bps);
			}
			printk( "Not reseted yet in Bp condition");
		}

		if(p_pulse_ox_conn == NULL)
		{
			// not currently connected to any pulse ox

			// check if we have a specific address to connect to
			if(have_target_pulse_ox_address == true)
			{
				// Have a target address - set filter
				//printk("Setting scan filter for target Pulse Ox address\n");
				bt_addr_le_t addr_le_pulseox;
				memset(&addr_le_pulseox, 0x00, sizeof(addr_le_pulseox));
				memcpy(addr_le_pulseox.a.val, target_pulse_ox_address, sizeof(target_pulse_ox_address));
				addr_le_pulseox.type = BT_ADDR_LE_PUBLIC;
				err = bt_scan_filter_add(BT_SCAN_FILTER_TYPE_ADDR, &addr_le_pulseox);
			}
			printk( "Not reseted yet in plus condition");
		}

		if(p_therm_conn == NULL)
		{
			// not currently connected to any thermometer

			// check if we have a specific address to connect to
			if(have_target_therm_address == true)
			{
				// Have a target address - set filter
				//printk("Setting scan filter for target Thermometer address\n");
				bt_addr_le_t addr_le_therm;
				memset(&addr_le_therm, 0x00, sizeof(addr_le_therm));
				memcpy(addr_le_therm.a.val, target_therm_address, sizeof(target_therm_address));
				addr_le_therm.type = BT_ADDR_LE_PUBLIC;
				err = bt_scan_filter_add(BT_SCAN_FILTER_TYPE_ADDR, &addr_le_therm);
			}
			printk( "Not reseted yet in threm condition");
		}

		err = bt_scan_filter_enable(BT_SCAN_ADDR_FILTER, false);
		if (err)
		{
			printk("Filters cannot be turned on (err %d)\n", err);
		}
		printk( "Not reseted yet in connetcion function");
	}
}


static int scan_start(void)
{
	int err;
	uint8_t chan_map[5] = { 0x55, 0x55, 0, 0, 0};

	err = bt_le_set_chan_map(chan_map);
	if (err)
	{
		printk("Bad Channel (err %d)\n", err);
	}

	// always refresh scan filters before starting scanning
	set_scan_filters();

	printk("Starting scanning\n");
	err = bt_scan_start(BT_SCAN_TYPE_SCAN_ACTIVE);

	if (err)
	{
		// error
		printk("Scanning failed to start (err %d)\n", err);

		//Scan Error
		uint16_t scan_error_payload = SCAN_ERROR_PAYLOAD;
    	send_message_data(SCAN_ERROR, &scan_error_payload, sizeof(scan_error_payload));
	}
	else
	{
		// success
		hub_scan_init_state = HUB_SCANNING;
	}

	return err;
}
  

we suspected it might be due to the stack overflow and we tried to increase the size however stil the issue sustained. we are not sure what the issue is since we could not find any debug error in the debugger kindly help us how to identify the issue related to the rest.

# disable DCDC

CONFIG_BOARD_ENABLE_DCDC_APP=n
CONFIG_BOARD_ENABLE_DCDC_NET=n
CONFIG_BOARD_ENABLE_DCDC_HV=n

# Enable the UART driver
# CONFIG_UART_ASYNC_API=y
CONFIG_UART_INTERRUPT_DRIVEN=y
CONFIG_NRFX_UARTE0=y
CONFIG_SERIAL=y
CONFIG_CONSOLE=y
CONFIG_UART_CONSOLE=y

# Enable the BLE stack with GATT Client configuration
CONFIG_BT=y
CONFIG_BT_CENTRAL=y
CONFIG_BT_SMP=y
CONFIG_BT_GATT_CLIENT=y
CONFIG_BT_MAX_CONN=20
CONFIG_BT_MAX_PAIRED=20
CONFIG_BT_CONN_CTX=y
CONFIG_BT_SMP_APP_PAIRING_ACCEPT=y


# Enable the BLE modules from NCS
CONFIG_BT_NUS_CLIENT=n
CONFIG_BT_SCAN=y
CONFIG_BT_SCAN_FILTER_ENABLE=y
CONFIG_BT_SCAN_UUID_CNT=3
CONFIG_BT_SCAN_ADDRESS_CNT=3
CONFIG_BT_GATT_DM=y
CONFIG_BT_GATT_DM_MAX_ATTRS=60
CONFIG_HEAP_MEM_POOL_SIZE=8192

# This example requires more stack
CONFIG_SYSTEM_WORKQUEUE_STACK_SIZE=16384
CONFIG_MAIN_STACK_SIZE= 4096

# Enable bonding
CONFIG_BT_SETTINGS=y
CONFIG_FLASH=y
CONFIG_FLASH_PAGE_LAYOUT=y
CONFIG_FLASH_MAP=y
CONFIG_NVS=y
CONFIG_SETTINGS=y

# Config logger
CONFIG_RTT_CONSOLE=y
CONFIG_LOG=y
CONFIG_USE_SEGGER_RTT=y
CONFIG_LOG_BACKEND_RTT=y
CONFIG_LOG_BACKEND_UART=n
CONFIG_LOG_PRINTK=n

CONFIG_ASSERT=y
CONFIG_BT_DEBUG_LOG=y

CONFIG_DK_LIBRARY=y

CONFIG_NO_OPTIMIZATIONS=y

#Debug Keys. MAKE SURE TO SET THESE TO n FOR PRODUCTION
CONFIG_BT_USE_DEBUG_KEYS=y
CONFIG_BT_STORE_DEBUG_KEYS=y

CONFIG_WATCHDOG=n

  • hi susheel, thank you for the mail, sorry for the delay, actually in using RTT viewer for viewing the debug messages. i face issues when using both the debugger and RTT viewer at the same time in visual studio code. i could not get the debugging messages. right now to check if the debugger breaks during reset. im sending a set of hex data to serial port to identify if the board resets. so everytime when the board resets i get the hex code in the serial port but still the break point does not breaks in the debug mode, and im using RTT viewer without debugging to check the debug messages and thats how i got the reason for reset that is soft reset. how can i now see both the debug message and debugger bith at same time to see when the break point breaks and see the reason.

  • hi susheel i misunderstood your message, actually you have mentioned to select the "optimize for debugging " option in the optimization level. However already the same option is selected during the project build. but still the code does not break during a reset. is there any other options to identify the exact cause of the reset

  • can you please check if you have CONFIG_RESET_ON_FATAL_ERROR=y in your prj.conf or in your .config file in your build folder.

    If that does not help, then please help me reproduce this on the DK and I will try to debug the soft reset and try to get the context of the reset.