I've been using the power profiler 2 kit for a couple of years and everything has been working fine - a few days ago I was promoted to do a firmware and version update (which I did) and suddenly everything is broken

the power profiler is now putting out 4.79v in ampere meter mode with no external power attached and in source mode, it is outputting 2.335v with the enable power output disabled. Setting 800mv output and enabling this raises the output to 2.897v out and the current measurements are now completely random with no correlation to load

is there a way to roll back to older firmware on this and undo the updates that have broken the device?

  • Hi Alethea,

    Could you try the following steps and see if the ppk2 works fine.?

    Disconnect all other nordic devices from the computer, turn off the kit and remove it from the plastic by removing the four screws. Insert USB to J2 (USB DATA/POWER). Short a connection between GND (e.g on the four pin power output header) to TP34 on the PPK2. While holding the jumper cable, turn the kit on and remove the jumper. Now the PPK2 should be in bootloader again.

    Open the PPK app, and select the device which is now just a serial number.

    When selecting it, the user will be prompted to program the application and it should now be working again.

    Best Regards,

    Swathy

  • thanks for that - I did the reset as you suggested and it reloaded 

    but it still seems to have the same strange behaviour

    when I first connect it to the mac or PC (I have tried two different computers and fresh installs) in source meter mode, with the supply voltage at the minimum (800mv) and the enable power output toggle set to OFF, there is 2.3v coming out of the vout pin, I can even light an LED with this 

    if i leave the enable power in the off position and raise the voltage in the software to max (5000mv) , we see 4.68V at the Vout pin even though the power output is switched off 

    If i switch the Enable Power supply to on, this rises above the 5V to 5.48V!

    If I set a source meter mode output of 3300mV we get out 3.207V when the out put is set to off or 3.932 when it is turned on

    testing this with a single 100Ohm resistor it shows a current draw of 17.43mA when Enable power is off and 18.89ma when on 

    so it seems the on / off has gone wrong and after the update, it is no longer able to turn off when in source meter mode

    and also, it seems perhaps the internal switching has malfunctioned as this is showing a significant difference in current regardless of which number we use

    3.207v / 100Ω should show 32.07ma or 3.932v should show 39.32ma vs 17.43mA and 18.89ma respectively - there was always a compensating voltage drop but never this sever - this indicates the ppk has an internal resistance of 100.8 Ω. when measuring 30ma which seems a lot higher than before the update


    Any ideas - can I roll back to the previous versions to test ? 