PPK Question

Hello,

We encounter the problem after upgraded the PPK software. There are different results when using PPK v1.0.0 and v1.1.0.

  1. Using PPK v1.0.0 image description

  2. Using PPK v1.1.0 image description

(The highlighted spikes are active by the timer timeout handler, and the others are active by the advertising.)

Which measurement is correct? and why?

Thanks.

Parents
  • Seeing the current bumps along the bottom line, it seems like the same power source is not used. Is the "Power select" switch in the same positition at both times? The new software is not doing anything different (when spike filtering is off) between the versions for this measurement.

Reply
  • Seeing the current bumps along the bottom line, it seems like the same power source is not used. Is the "Power select" switch in the same positition at both times? The new software is not doing anything different (when spike filtering is off) between the versions for this measurement.