I’ve provided radiated immunity test for our devise based on the nRF52832.
The test has been provided by electromagnetic pulse with a power of 10V/m ,duration 1sec and the same test with a duration 3sec, the frequency range from 2GHz to 2.7GHz with a frequency grid 1% between each hopping.
During the test in the frequency range from 2.4GHz to 2.7GHz wasn’t observed any interference of the device .
But in the lower frequency range from 2GHz to 2.345GHz with a duration pulse of 1s, were observed difficulty in signal transmission.
For the pulse duration of 3s, the connection between device and computer was completely interrupted and stop pairing .
The problematic frequencies are 2.165,2.231,2.253,2.298,2.345GHz.
This test has been provided In accordance with the standard for medical home use , the minimum pulse duration must be at least 1s.
Have to be taken into consideration that harmful frequencies are not in range of the BLE frequency range. I want to know what affects to processor ,at low frequencies, stop to work correctly?