ppk gives wrong value.

Hi,

I am using ppk with nrf52832 dk to measure the current consumption of my board and it giving different values at different times. for two days it gave 250uA and tested it with 5/6 different boards and all were in same range. but after 4 days when try again, with same sw and hw without any change, it gives 34-40 mA.

My connections,

DUT Switch : external

power supply switch : reg  (at 3.6 V)

COM switch : dk

nrf ppk sw 2.6.1 and 2.5 on ubuntu and windows.

I have connected a 10k and 100k resistors as load and measured the consumption and it was giving correct value.

I have tested the power consumption of all these same boards  with gold uCurrent  and it gives the values as 800uA consistently across all the devices.

How can i make sure that my ppk is working fine? what are all the ways to trouble-shoot ppk ?

Parents Reply Children
  • i think that device got damaged, any way i got a new ppk and tested the board. i think it gives proper values, i have measured by keeping the switch in reg position.

    i will test it one more time with sw in dk position and try to follow taht.

    Also it seems there is some bug in your gui, while increasing the regulator voltage to maximum though the ui, it shows unusual activities and unrealistic values. once you slightly move it from extreme end of 3.6, it comes to normal values.