The current measurement values ​​of the Fluke equipment and the PPK2 are very different.

After purchasing the PPK2, I connected the Fluke287 and PPK2 Ampere meter in series to check whether the measured values ​​matched the Fluke equipment. 

The Fluke287 consistently checked 37uA, but the PPK checked about 3uA, so the difference is too big. What is the reason?
Parents Reply
  • This will not work. Read through the user guide for the PPK2, this shows all ways to connect loads that will work and don't try other options as they are not supported.

    You need to connect the Power Supply to VIN and GND and then you connect the 1M resistor from VOUT to GND if you want to test the current through the resistor..