PPK2 Measurement Issue

Hi,


I have two PPK2 units, and under the exact same test environment and PPK2 settings, simply swapping the PPK2 results in completely different measurement results.
Which PPK2 should I trust, and what should I do with the other one?
Thanks!

PPK2-1

PPK2-2

Parents Reply Children
  • Hi,
    I ran a few more comparisons and selected the two images that show the most significant differences.


    From these images, it seems clear that the two PPK2 units have a noticeable discrepancy in their ability to capture peak current.

    However, I don’t have any other instruments on hand to verify which one is accurate.
    Which PPK2 should I trust? Or do you have any other suggestions?
    Thanks!

  • This can be also related to the PP2 will have challenges sampling the nPM2100 input current. The sampling rate of the PPK2 is not fast enough to really record the current pulse caused by the boost converter operating in hysteretic mode of the nPM2100. The dynamic range needed for this measurement causes difficulties since the current peaks can be in the tens of mA range, but the base current is <1uA. So it needs autorange to record this.

    Adding bigger bulk capacitor in parallel with the nPM2100 input would reduce the current peaks and make them longer. This should help at least reducing the error seen.

    Kenneth