Hi DevZone,

I am using the nRF52840 to interface an ADS1298 (ECG Chip) via SPI.

I want to test if the communication between the devices are working by:

    -  Sending a RREG (Read From Register) opcode from the nRF52840 to the ADS1298

    -  And then read its ID Register.

The RREG command is two bytes long, and for multi-byte commands the following needs to be done:

    1.  Send the first byte and then wait for 4 * tCLK  before sending the next byte.

    2. The CS (or SS) pin needs to be held low during the entire session for both transfer and receive.

I am trying to test this with the example provided in the nRF5 SDK 15.3 by using this function:

    -  nrf_drv_spi_transfer(&spi, m_tx_buf, m_length, m_rx_buf, m_length);

My problem is that:

    -  I cannot insert a delay between the bytes when using the nrf_drv_spi_tranfer() for multiple byte transfers

    -  If i use separate calls of the nrf_drv_spi_transfer() for each byte, then the CS pin is not held low.

Is there any way I can use this driver to send multiple bytes with a fixed delay between them, without the CS pin going high during the session?

Thank you for reading

Br. Casper

Parents
  • You don't necessarily need to add a delay, just to conform to the AFE requirements.

    Since transfers are only a few bytes, typically the slower clock speed is not an issue. Other issues tend to be more problematic; as an example the internal AFE clock has drift with respect to external timing sources, such as the crystal-clocks used further down the chain in ECG analytics. This has to be handled without introducing artifacts in the ECG stream.

    Using an external crystal clock is only one option to minimise this, but as a tip note that without a clock the ADS129x AFE is unavailable via SPI. So if the clock select pin is set to external clock mode the SPI interface will simply not work until the external clock is present. A classic mistake is to use the AFE output pin to select the clock source, which if defaulted to external (via a pull-up) means first an external clock has to be supplied to allow SPI transfers, then the pin programmed low for internal clock via SPI, then the external clock can be removed again.

    To manually control the CS pin, set the pin to not used; this is the nrf example, but may differ slightly depending on which library versions you are using

Reply
  • You don't necessarily need to add a delay, just to conform to the AFE requirements.

    Since transfers are only a few bytes, typically the slower clock speed is not an issue. Other issues tend to be more problematic; as an example the internal AFE clock has drift with respect to external timing sources, such as the crystal-clocks used further down the chain in ECG analytics. This has to be handled without introducing artifacts in the ECG stream.

    Using an external crystal clock is only one option to minimise this, but as a tip note that without a clock the ADS129x AFE is unavailable via SPI. So if the clock select pin is set to external clock mode the SPI interface will simply not work until the external clock is present. A classic mistake is to use the AFE output pin to select the clock source, which if defaulted to external (via a pull-up) means first an external clock has to be supplied to allow SPI transfers, then the pin programmed low for internal clock via SPI, then the external clock can be removed again.

    To manually control the CS pin, set the pin to not used; this is the nrf example, but may differ slightly depending on which library versions you are using